[Test-Equipment] Resistor measurement
Richard Knoppow
dickburk at ix.netcom.com
Wed Jul 6 03:54:04 EDT 2011
----- Original Message -----
From: "Fuqua, Bill L" <wlfuqu00 at uky.edu>
To: "Discussion of Electronic Test Equipment"
<test-equipment at mailman.qth.net>
Sent: Tuesday, July 05, 2011 10:33 PM
Subject: Re: [Test-Equipment] Resistor measurement
> If you had 4 of them you could make a bridge and use a
> sensitive meter
> to check the balance. If any one of them was off there
> would be an error or imbalance and
> it is unlikely all of them would have the same error.
> 73
> Bill wa4lav
>
If you have two of something you can compare them to
the degree of precision of the measuring device. There is a
difference between accuracy and precision. An instrument may
have a high degree of precision, that is, be able to detect
and read very small differences repeatably and reliably, but
not be accurate. However even an uncalibrated instrument can
make a comparison reading within the limits of its
precision. An instrument can not be calibrated to an
_accuracy_ greater than its precision, however it can make
differential readings up to that limit.
In fact, the substitution method is a good one for
calibration of devices such as resistors since it eliminates
the tolerance of all components other than the method of
reading out the differences.
The usual standard for resistance calibration is a
Thomas Ohm cell. These are extremely precise one ohm
resistors of known error. They can be certified by NIST and
used to calibrate working standards at a DC standards lab.
The less accurate and stable standard is called a Rosa
resistor or Rosa bottle because they resemple bottles. These
can also be certified by NIST but are usualy used as working
standards after being calibrated against a certified Thomas
Ohm or group of them. Its been a long time since I
maintained a DC standards lab for a resistor manufacturer. I
think our master bridge was an Otto Wolf bridge which was in
an oil bath to maintain its temperture constant. We had a
group of about a dozen Thomas Ohms, which were cycled
through (at that time) NBS. Precise resisance measurements
can be difficult despite being DC. Especially difficult is
the measuement of very small resistances since the
difference in potential of the metals at the terminals can
affect the readings. We had a very good Kelvin Double Bridge
to measure fractions of an ohm. Oh well, another life...
--
Richard Knoppow
Los Angeles
WB6KBL
dickburk at ix.netcom.com
More information about the Test-Equipment
mailing list