[Test-Equipment] HP 8569 Info On-line ????

Skip Frolik [email protected]
Sun, 22 Feb 2004 11:55:48 -0600


Marco,
Thanks a bunch for all the info.
Sounds like a keeper.
Appreciate you taking the time to
respond with such a detailed e-mail.
Thanks again.
Skip
WB4GMQ
-----Original Message-----
From: Marco Bruno [mailto:[email protected]]
Sent: Sunday, February 22, 2004 02:26 AM
To: [email protected]; Skip Frolik
Subject: Re: [Test-Equipment] HP 8569 Info On-line ????


At 00.13 22/02/2004, you wrote:
>  Whether this unit is good
>or not remains to be seen.

Hello Skip and all,

I don't know about your particular unit, but I'm used to recondition and
sell 8569B's, and it's a very good analyzer. There is not much online about
it. Good sensitivity, narrow filters down to 100 Hz, clean LO, good level
accuracy and flatness. NOT portable, HI. It drives a plotter directly on
the HPIB (7470A or 7475A -like). May drive the 8444A-059 tracking generator
up to 1.5 (1.7...) GHz. Built-in preselector from 1.8 to 22 GHz; direct
support of external waveguide mixers up to 115 GHz. Those for 18-26 and
26-40 GHz bands are easy to find (11971K and 11971A). Has a
first-generation digital display, unfortunately without markers, but with
two memories and peak-hold function. The display may operate in analogic
mode, very good to see pulses (radar or your Wi-Fi link). The LO is not
synthesized, but frequency accuracy for a well maintained machine is around
a couple of MHz at UHF, rising to abt 5 MHz a 10 GHz. Get the service
manuals (3 binders), definitively worth having. Easy to repair, very good
accessibility to inside. Classical problems: stuck (or defective) input
relay, easy to replace. Quite long to calibrate if somebody messed up the
trimmers... expecially the band level adjust trimmer bank. Delicate front
panel mechanics, like all the HP 8558B-8559A-8565A... rotary switch wipers
and clutches are quite difficult to fix if damaged. If you need performance
data I may scan the relevant pages from the HP catalog.

Moe or less it is similar to the Tektronix 492P opt. 1-2-3, but I prefer
the 8569B. More bulky, but easier to operate, and far better LO. The 8565A
is very similar, but has a storage CRT insted of the digital memory; HP
says it is less sensitive too, but I never verified it is true, or was a
commercial issue - the RF section is so similar...

Vy 73s - Marco IK1ODO