[Test-Equipment] Cosel K150A-12 Power Supply
Fred Olsen
[email protected]
Mon, 26 May 2003 23:23:54 -0500
Arthur, as has been pointed out those terminals are indeed for remote
voltage regulation sensing. There are a couple of fine points for
implementing remote sensing which should be mentioned.
In general, use twisted pair for your load lines. 16 Ga. is sufficient
for this supply. Use shielded twisted pair for the sense lines.
There's a note regarding gauge on the below-referenced data sheet but
common 20 or 22 Ga. STP instrumentation cable will work fine in most
cases. The 100% shield foil stuff with a drain wire works the best; the
sense circuit must be kept quiet. Ground the shield at one end only,
usually the PS. Note that the sense lines should use a Kelvin
connection at the load. Don't cheat and tie them to the load lines
further back. Avoid switches, relays, and connectors in the sense
circuit if at all possible.
This next is so obvious that "it goes without saying". Having seen it,
I'll say it. When using remote sensing the local sensing jumpers
between the V and S terminals must be removed. Both of them. (Saw that
too.) Without remote sensing they must be in place.
In any application of remote sensing the following is important but
seldom mentioned. Make certain that your power supply has sensing
protection diodes at its output. If the remote sense circuit suffers an
open for any reason the supply will drive to maximum possible output
voltage. If your load can stand whatever that may be, good for it. If
not, a cheap pair of 4000-series diodes will save your bacon by limiting
the output to one diode drop above the set point. Check the schematic,
if the Malaysians gave you one, or visually inspect for the presence of
a pair of diodes, one between each set of V and S terminals. Or they
might be the 'last thing' on the board before the output. I suggest
adding them if they aren't there. They are arranged thus: anode to S- /
cathode to V-, and anode to V+ / cathode to S+.
Sometimes resistors, on the order of 100 Ohms, are found as sensing
protection. Diodes are preferable.
Here's a handy poop sheet on remote sensing, from 'the company formerly
known as HP'.
http://www.measurement.tm.agilent.com/hints/5965-8239E/hint01.shtml
Those who don't consider themselves expert in power supply application
would do well to find a copy of HP App Note 90B, "DC Power Supply
Handbook". An oldie but goodie.
Best of luck with the project,
Fred
--
<><
--
Outgoing checked by Norton AV