[R-390] R390A AGC troubleshooting procedure
Charles Steinmetz
csteinmetz at yandex.com
Mon Jan 6 15:21:27 EST 2014
I wrote:
>[TB102] Terminal 4 is the AGC line feeding the RF and IF circuits
>and should read essentially infinite (>> 1M ohm).
Dave wrote:
>No, there's a voltage divider in the RF deck. R201 and R234, on
>Figure 5-18, page 39 in Y2K-R3.
>Resistance to ground will be 1.8M . When you unplug P108 off the RF
>deck, terminal 4 should go open.
Right you are, Dave -- thanks! Here's a further corrected version of
the procedure:
R390A -- AGC troubleshooting procedure:
Throughout this entire procedure, the "FUNCTION" switch should be set to "AGC."
First, make sure there is a jumper installed between TB102, Terminals
3 and 4. If not, install one and see how the radio works now.
Set the AGC to "MED" and tune the radio to a good, strong, local
signal (like a strong AM broadcast station). Measure the DC voltage
at TB102, Terminals 3 and 4 with a high impedance meter (VTVM, DVM,
or scope, with an input resistance of >= 1M ohm; not a VOM). It
should be significantly negative, -10v or more. If it is, you have
no gross AGC problem and the fault lies elsewhere. But if the
voltage is only weakly negative, or zero, you have an AGC problem. If so:
Turn off the radio, and pull the plug. Set the AGC time constant to
"MED." Remove the jumper between TB102, Terminals 3 and 4. Measure
the resistance to ground from each of these Terminals. Terminal 3
looks back into the AGC detector, and should read in the neighborhood
of 500k ohm due to R545, R546, and R547. Terminal 4 is the AGC line
feeding the RF and IF circuits and should read approximately 1.8M ohm
due to R201 and R234. If you unplug P108 from the RF deck, Terminal
4 should then read essentially infinite (>10M ohm). If you have
gotten to this point, one or the other of these Terminals will
probably show a much lower resistance to ground than this. Trace the
circuit to find the leaky component(s).
If Terminal 3 reads less than ~500k ohm, the usual suspects are C551,
C548, C547, C545, and C544.
If Terminal 4 reads significantly less than ~1.8M ohm, the usual
suspects are any of the several dozen bypass caps on the AGC line in
the IF and RF sections, or leakage to ground in one or more of the
mechanical filters. It is also possible that the sector of the
"FUNCTION" switch that shorts Terminal 4 to ground when the switch is
set to "MGC" or "STAND BY" is mis-timed, broken, or dirty, but this
is unlikely.
If, on the other hand, the resistance readings are OK, suspect V508,
V509A, and associated circuitry (Z503 and C546, especially).
The R390 is very similar, although the part numbers are different.
Best regards,
Charles
More information about the R-390
mailing list