I like your writing; it's clear and easy to follow.
> Terminal 4 is the AGC line feeding the RF and IF circuits and should read essentially infinite (>> 1M ohm).
No, there's a voltage divider in the RF deck. R201 and R234, on Figure 5-18, page 39 in Y2K-R3.
Resistance to ground will be 1.8M . When you unplug P108 off the RF deck, terminal 4 should go open.
Dave Wise
-----Original Message-----
From: r-390-bounces at mailman.qth.net [mailto:r-390-bounces at mailman.qth.net] On Behalf Of Charles Steinmetz
Sent: Tuesday, December 31, 2013 1:16 PM
To: 390 list
Subject: [R-390] R390A AGC troubleshooting procedure
I posted this on another thread yesterday -- I'm reposting with some
additions and so it has the correct "Subject:" header.
R390A -- AGC troubleshooting procedure:
Throughout this entire procedure, the "FUNCTION" switch should be set to "AGC."
First, make sure there is a jumper installed between TB102, Terminals
3 and 4. If not, install one and see how the radio works now.
Set the AGC to "MED" and tune the radio to a good, strong, local
signal (like a strong AM broadcast station). Measure the DC voltage
at TB102, Terminals 3 and 4 with a high impedance meter (VTVM, DVM,
or scope, with an input resistance of >= 1M ohm; not a VOM). It
should be significantly negative, -10v or more. If it is, you have
no gross AGC problem and the fault lies elsewhere. But if the
voltage is only weakly negative, or zero, you have an AGC problem. If so:
Turn off the radio, and pull the plug. Set the AGC time constant to
"MED." Remove the jumper between TB102, Terminals 3 and 4. Measure
the resistance to ground from each of these Terminals. Terminal 3
looks back into the AGC detector, and should read in the neighborhood
of 500k ohm due to R545, R546, and R547. Terminal 4 is the AGC line
feeding the RF and IF circuits and should read essentially infinite
(>> 1M ohm). If you have gotten to this point, one or the other of
these Terminals will probably show a much lower resistance to ground
than this. Trace the circuit to find the leaky component(s).
If Terminal 3 reads less than ~500k ohm, the usual suspects are C551,
C548, C547, C545, and C544.
If Terminal 4 reads less than 1M ohm, the usual suspects are any of
the several dozen bypass caps on the AGC line in the IF and RF
sections, or possibly leakage to ground in one or more of the
mechanical filters. It is also possible that the sector of the
"FUNCTION" switch that shorts Terminal 4 to ground when the switch is
set to "MGC" or "STAND BY" is mis-timed, broken, or dirty, but this
is very unlikely.
If, on the other hand, the resistance readings are OK, suspect V508,
V509A, and associated circuitry (Z503 and C546, especially).
The R390 is very similar, although the part numbers are different.
Best regards,
Charles
______________________________________________________________
R-390 mailing list
Home: http://mailman.qth.net/mailman/listinfo/r-390
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:R-390 at mailman.qth.net
This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html
More information about the R-390
mailing list