[R-390] Distortion on MGC mode and another question on sensitivity
n4buq at knology.net
n4buq at knology.net
Thu Mar 15 09:41:08 EST 2007
Thanks for the replies about the MGC distortion. I was thinking someone
had posted a while back that the R390* radios should not overload on strong
signals, period, but I suppose that's only in AGC mode.
Another question: When I'm testing for sensitivity, I'm looking for the
signal level that will result in -7V on the Diode Load; however, when I
connect the signal generator, there is atmospheric noise that is introduced
before I inject any signal which results in a few volts on the Diode Load.
Is it correct to increase the generator's output level to -7V total and
read the sensitivity from there, or should the signal generator result in a
7V increase over the existing noise? In other words, if I'm getting -3V in
atmospheric noise, then do I take my sensitivity reading at -7V or -10V?
Part of this problem is the rigged-up mechanism I have for connecting the
generator to the receiver. I'm in the process of making a housing for the
attenuator and feeding everything with coax thus eliminating the
little "antennae" I introduce with the current setup.
I assume the -7V reading is accurate (or at least more accurate) if the
test is done in an RFI-proof environment with a perfect signal generator
that doesn't leak. Lacking those two things, what is the proper way to
measure this?
Thanks again,
Barry