>From: Barry Hauser <barry at hausernet.com>
>To: Kenneth Arthur Crips <CRIPS01 at MSN.COM>, r-390 at mailman.qth.net,bernie
>nicholson
and there can
>be errors in the tube charts. The manufacturers did not necessarily
>calculate the correct minimum reading for every number. Many of the
>entries were probably determined through extrapolation/interpolation, if
>that. (Or they cheated and used a conversion factor on the values from
>some other tube tester.)
I can tell you folks how I recall we college co-ops did it back at Hickok...
When the tube manufacturers introduced new tube types, 3-5 tubes were
normally received along with the new design specs. Each of these new tubes
was set up on a super lab tester at exactly the published specs to determine
precisely how each one compared to the published specs. For example, one
tube might have 104% of the rated Gm while another might have only 97% of
the rated Gm.
I believe Hickok's testers usually had a fixed plate voltage but some could
be set within limits. It was the bias that was adjustable with a front
panel control. When one of these new tubes was set up in a particular tube
tester, the bias was adjusted to approximate the plate current that was used
in the manufacturers determination of Gm. When you had about the right
plate curent the shunt control was adjusted so that a tube with 100% of a
new tube Gm would read exactly 2/3 full scale on a fail-?-pass type of
scale. The bias and shunt readings that you set up would be what appeared
on the next roll chart. If the tester had calibrated Gm scales, the bias
would be adjusted until the meter read exactly what the lab standard read
for that tube.
Note that the tube would only have to drop to about 70% of its new tube
value before it reached the bottom of the "?" area and start to be rejected.
That why a lot of tubes that test in the fail or reject part of the scale
still work in equipment. Remember, tube manufacturers wanted to sell tubes
and it was better to err on the side of conservatism (better to throw away a
still usable tube than to keep a bad one).
When I was doing this back in the middle 60s, there were only about 3-4
models that had their roll charts upgrated with the new tube types. There
was also that cardmatic one that got the new types. If you wanted data on a
tube for an older model that didn't have it on its roll chart all you had to
do was call customer service. I remember dragging those oldies out of the
'morgue' and getting the settings for a customer or two. A new cardmatic
model was also being develolped for the military but I don't think that ever
got pass the prototype stage. I remember doing the layout on the power
supply pc boards for those. By that time tubes were on their way out.
Introducing new tester models was a nightmare...you had to go and verify
every tube type with the new tester settings. I spent many an hour doing
that on their 'mustang' model. Kind of boring but lots of overtime.
enough rambling....
pete, KH6GRT