[Heathkit] FW: TT-1A Issues - MORE DETAILS
Gary H. Harmon, Jr.
gharmon at idworld.net
Wed Oct 4 21:48:46 EDT 2017
From: Ron Eisenbrey [mailto:ab5wg at satx.rr.com]
Sent: Wednesday, October 04, 2017 7:02 PM
To: 'Gary H. Harmon, Jr.'
Subject: RE: TT-1A Issues
6V6 tube will not test in the TT-1A. Multiple tubes, all do not test and
are known to be good. There is no response from the tester when the gm test
switch is operated. Other 8 pin tubes test good in the same socket, only
the 6V6 will not respond. (Perhaps a wrong setting in the tube data?)
12AU7 tube always measures extremely low with the switch settings given in
the tube data. If I decrease the bias, tube will test good. This happens
with any 12AU7, all test fine in other testers. (Again, I suspect something
is wrong in the tube data settings). Bias is calibrated, as are all other
internal calibrations on the TT-1A. All other 9 pin tubes test good in the
same socket, just the 12AU7 always tests low.