[GPS_Standard] Self Test experiment with strange result
Dimitri.p
dimitri at dotp.com
Fri Nov 6 16:47:13 EST 2015
Now that I seem to have "Average Frequency
Offset" seperated from "Average frequency offset
ERROR" in my head I'll stop banging it against the wall :)
Thanks !
Dimitri
At 01:06 PM 11/5/2015, Peter Mulhare wrote:
>Hi Dimitri
>
>What Montrol is giving you are Average Frequency
>Offset, Standard Deviation, and Minimum Offset
>and Maximum Offset. These are done in the
>Current Cycle, the Previous Cycle, and the Accumulated Totals.
>
>If you reset Montrol completely ie quit the
>program when the GPSDO is running and is stable
>and locked, then reboot Montrol, and watch the
>Average offset in the yellow Column for a while.
>
>Immediately Montrol starts there are no averages
>and the Current Cycle Offset you will see are
>the actual offset figures, the Accumulated value
>will also exactly follow for a while but begin
>to change less as the number of totals build up
>over time. Once there are a reasonable amount
>of completed cycles you will start to see the
>average offset develop of all the cycles from Montrol switch on.
>
>You will notice that over a much longer time
>that the figures in the Blue Column (Accumulated
>Totals) will get closer and closer to the
>Statistical Offset of the GPSDO. Because the
>GPSDO is a Frequency Locked Loop, not a Phase
>Locked Loop, the actual GPSDO frequency will
>change slightly over time, this is because of
>jitter in the 1pps from the GPS, due to various
>factors, and also the tendency for the crystal
>to wander slightly in Frequency, all these
>errors accumulate over time and Montrol is
>showing you, what the statistical error of the
>GPSDO is (from a Statistical Bell Curve) at any
>given time. It is not giving you at any time the Actual Frequency!
>
>Montrol is showing you 3 slightly different ways
>to express the Statistical version of all the
>offsets to date since Montrol was switched on.
>
>This is a simplistic explanation but I think
>that it will explain it correctly for you.
>
>With regard to your last post that is correct it
>is not able to read any better than 6.25 x
>10-9. But remember that is the max offset above
>& min offset below, which I would expect to be
>greater than that. Mine is typical around 3.12
>x 10-8. It is not an average. And remember
>that 6.25 x 10-9 is 6.25 parts in 1GHz!
>
>With regard to noise on the EFC line, I took the
>precaution of using shielded audio cable and
>only earthed the shield at the electronics end
>and not at the Oscillator end. I also shielded
>the 1pps line from the GPS, and earthed that at
>the GPS end only. Im using a 12Volt CIC
>ST2145A double oven Oscillator, and an Adafruit
>GPS on a small breakout board. The GPS has a
>rated jitter of +10 nS rms. My system is
>working as I would expect it to. I have
>replaced the oscillator 3 times, twice with
>Isotemp OCXO131-100, one of which went faulty,
>and finally with the double oven CIC
>ST2145A. To be honest there has not been much
>difference statistically between all 3
>OCXOs. I built the unit nearly 4 years ago.
>
>Pete Mulhare
>ZL2iK
>
>
>
>-----Original Message-----
>From: GPS_Standard
>[mailto:gps_standard-bounces at mailman.qth.net] On Behalf Of Dimitri.p
>Sent: Friday, November 6, 2015 08:48
>To: gps_standard at mailman.qth.net
>Subject: Re: [GPS_Standard] Self Test experiment with strange result
>
>Thanks for the reply.
>Although it justifies the existence of "0x6800" , if we take
>160000001 and divide it by 16 it results
>in 10000000.0625 Mhz. Now unless I am still
>missing something any change less that .0625 cannot be "seen" by the FLL.
>So my realization is in a "best" case scenario when the FLL displays
>6800 the actual oscillator frequency is
>somewhere either exactly 10000000.00000000
>or between plus or minus 0.06249 MHz
>
>And, correct me if I'm wrong, if indeed the
>Min Max display in Montrol is the ppm error the
>way you described, it will never be able to
>measure anything better than plus or minus 6.25e-9
>
>
>Dimitri
>
More information about the GPS_Standard
mailing list