[Elecraft] K3; VM calibration and Voltage drops resolved
Richard Fjeld
rpfjeld at embarqmail.com
Sat Dec 29 10:16:40 EST 2012
This is as short as I can make it. When I first noticed the difference in the K3 meter reading compared to my DVM, I wondered if I needed to calibrate something, and I wondered if the K3 meter was linear. The answer is; no, and can't tell.
K3 volt meter at key up: 13.7 V DVM at PS: 13.71 V
K3 volt meter at key down: 12.8 V DVM at PS: 13.68* V
K3 12 volt out jack w/key up: DVM 13.01V
K3 12 volt out jack w/key down: DVM 12.28* V
(at first glance, this leaves some questions)
In summary;
--I measure no Vd through my homebrew distribution blocks.
--I calculate no Vd through the Power Poles. (Difficult to test through.)
--There is minor drop in power wires.
--K3 12V out jack reads 1.4 V lower than PS voltage at key down. Two diodes in circuit to the 12V out jack explain that drop.
--The 12V out jack cannot be used to measure voltage drop from PS.
--Unsure of the VM circuit test point for linear comparison. No external access.
Thanks to those who contributed,
Rich, n0ce