[Elecraft] K2- Is the Power display really representative of PEP?
W3FPR - Don Wilhelm
w3fpr at earthlink.net
Sun Nov 7 11:37:15 EST 2004
Robin,
The base K2 power display is a simple diode detector (an RF probe is
similar), and it will be reasonably accurate if and only if the load
inpedance is 50 ohms resistive. Your antenna may be something quite
different. You might say that the base K2 power output indication shows the
power that would be present if the antenna load were actually 50 ohms
resistive.
If you have the KAT2 or KAT100 or KPA100 installed, they contain a true
wattmeter, but these must be properly calibrated to provide a good reading.
In general, wattmeters can be trusted to only 10 to 20% of their full scale
reading - so if you are reading on a 100 watt scale, the actual power may be
off by as much as 10 to 20 watts!!!
Use the lowest scale that will handle the power being measured. Actually
some are better than that, but should not be trusted until they are properly
calibrated. I calibrate (or verify) wattmeters by connecting a 50 ohm dummy
load and measuring the RF voltage across that load with my 'scope (an RF
Probe could also be used) - then do the math to determine the power. When
using the 'scope, the formula I use is Vp-p ^2/400 and for an RF Probe that
reads RMS voltage the formula would be V^2/50 (these formulas are only for a
50 ohm load). Check the dummy load beforehand with an antenna analyzer to
be certain it is a good 50 ohm resistor at the frequency you are using for
the calibration. In other words, trust nothing for accuracy until it is
verified.
73,
Don W3FPR
----- Original Message -----
I've probabaly seen this discussed before, but can't remember ...
I have noticed that the power output on my K2 differs from measures PEP
going to the antenna. Mostly it displays between 14.8and 15.2.
However, when measuring with a wattmeter and dummy load I notice that on 18m
I get 18W PEP, and on 15m down to 12W. On 10m it's at 8W.
So my question is, is the power level that you seen on the display really
just a scale for adjusting power, or is it at true reading. I'm starting to
wonder if my Watt meter is inaccurate (a Yaesu 60 model) ?