With my newly finished XG1 I am getting a big difference in the Cal S Hi =
and
Lo settings when calibrating my K2 sn 03029 at 50uV per the XG1 manual =
(Rev
C) compared to the K2 manual (Rev D) method.
This K2 has kat2, knb2 and kaf2 installed. (Kat2 set at ATU CalS =
during
measurements.) The mds measured was -140dbm. (1.87v & 0.039v on ext =
speaker
jack). All XG1 assembly resistor value checks were ok.
The initial settings (preamp ON)per the K2 manual page 46 (pin5 U2 =3D =
3.7V)
were CalS Hi =3D 012(bar 10 on), Lo =3D 187(bar 1 flickering). With =
these
settings and the XG1 set at 50uV (preamp ON) bar 10 was solidly on. =
Turning
off the Preamp showed bar 9 on (S9+30?) Turning on the Attenuator =
(-10db)
brought it back to bar 6 on (S9 indication)
If CalS Hi and Lo were adjusted per the XG1 manual (preamp ON) the =
settings
were 026 (bar 6 on) and 181 (bar 1 flickering) respectively but then =
with
the RF Gain turned ccw and no signal the highest bar to light is bar 5.
Am I missing something, or does the XG1 have a >50uV output or do I just =
use
the original CalS settings and know I need to turn the attenuator on to =
get
a comparative S9 reading?
Thoughts and suggestions would be appreciated
Nigel
-----------------
ZL2DF
New Plymouth
New Zealand
[email protected]
K2 #03029