[Elecraft] K2 Freq-Calibrate
Masleid, Michael A.
[email protected]
Mon Jul 21 17:04:00 2003
I was wondering how accurately I could calibrate the K2 using the
method proposed by Steve Gibbs GU3MBS last December. I set up
an OCXO controlled PLL to generate 4 MHz, verified that it was
within 0.02 ppm of WWVB, and used that to trigger a scope. I put
a probe near C22 and turned up the gain enough to see the 4 MHz =
oscillator,
then adjusted C22 so that drift was less than a fraction of a cycle per
second. It takes patience, but it sets up the 4 MHz oscillator within
0.1 ppm. I then ran CAL PLL and CAL FIL and checked calibration on
several bands by zero beating with SPOT.
Calibration is off by about 3 ppm, for instance, WWV at 10 MHz comes in
at 9999.97 KHz. What is happening?
I did this with #3430, with the PLL temperature fix. At page 45 in the =
manual,
the test of the frequency counter, I got 3999.98 KHz. This means that =
the counter
is 5 ppm low, give or take 2.5 ppm. That doesn't change with C22, since =
the
4 MHz oscillator is the reference for the counter. This counter being =
off will
cause CAL PLL and CAL FIL to set everything at too high a frequency, so
stations will seem to be somewhere between 2.5 and 7.5 ppm low - in my
case, 3 ppm.
If I set C22 so that the counter is on the money, the 4 MHz oscillator =
will be off
a bit. This will throw the internal "birdies" off frequency a little, =
but that's OK. Is
this normal?
I'm sure I can set C22 by the book - but using a reference oscillator =
and scope
looked like it would be fun to try.
Michael AB9GV