For the test points that should read >100k Ohms, I get different readings
depending on what scale I am reading on the ohm meter. I am using a Micronta
meter I purchased at Radio Shack several years ago that has the following Ohm
meter ranges 200, 2k, 200k 2m 20m.. For the test points in question, I get
different readings on the 200k, 2m and 20m range.. All readings are OK on the
highest range, but may be close in the 200k range..
The meter reads purely resisitive loads (resistors) fine on all ranges..
=====================================================
Digital voltmeters measure resistance by forcing a constant current across
the test leads and measuring the voltage across those leads. For example by
Ohm's Law if the meter forces 1.000ma with a 1000 ohm resistor across the
test leads, the voltage will read 1.000 volt. The current forced changes when
you change ranges. Also most DVM's limit the voltage on some ranges below 0.5
volt so semiconductor junctions will not be biased on. While other ranges
allow about 2 or 3 volts maximum. This is why the resistance is different on
some ranges. When a diode, IC or transistor is biased on with the higher
voltage the current flowing in the semiconductor will upset the reading. Some
meters (Fluke is one) have a little diode symbol on the ohm range switch to
let you know it will bias on a semiconductor. When the manual asks you
measure >100k any value larger than 100k is OK. I would measure this on the
200k range because the lower ranges will just read overrange. If the meter
reads overrange on the 200k range I know the reading is over 200k. Sometimes
you will have to wait a second or two for any capacitors to charge before the
reading is stable.
BTW the ranges that allow the biasing on semiconductors can be used for
matching diodes. Remember the meter is measuring the voltage across the test
leads so the number is the forward voltage drop of the diode or Vf. Just
select diodes that have the same reading.
Don Brown
KD5NDB
--part1_18b.1486997.29620c9f_boundary
Content-Type: text/html; charset="US-ASCII"
Content-Transfer-Encoding: 7bit
<HTML><FONT FACE=arial,helvetica>In a message dated 12/31/2001 10:36:07 AM Central Standard Time, [email protected] writes:<BR>
<BR>
<BR>
For the test points that should read >100k Ohms, I get different readings depending on what scale I am reading on the ohm meter. I am using a Micronta meter I purchased at Radio Shack several years ago that has the following Ohm meter ranges 200, 2k, 200k 2m 20m.. For the test points in question, I get different readings on the 200k, 2m and 20m range.. All readings are OK on the highest range, but may be close in the 200k range..<FONT SIZE=3><BR>
<BR>
</FONT><FONT COLOR="#000000" style="BACKGROUND-COLOR: #ffffff" SIZE=2 FAMILY="SANSSERIF" FACE="Arial" LANG="0">The meter reads purely resisitive loads (resistors) fine on all ranges.. </FONT><FONT COLOR="#000000" style="BACKGROUND-COLOR: #ffffff" SIZE=3 FAMILY="SANSSERIF" FACE="Arial" LANG="0"><BR>
<BR>
</FONT><FONT COLOR="#000000" style="BACKGROUND-COLOR: #ffffff" SIZE=2 FAMILY="SANSSERIF" FACE="Arial" LANG="0">=====================================================<BR>
<BR>
Digital voltmeters measure resistance by forcing a constant current across the test leads and measuring the voltage across those leads. For example by Ohm's Law if the meter forces 1.000ma with a 1000 ohm resistor across the test leads, the voltage will read 1.000 volt. The current forced changes when you change ranges. Also most DVM's limit the voltage on some ranges below 0.5 volt so semiconductor junctions will not be biased on. While other ranges allow about 2 or 3 volts maximum. This is why the resistance is different on some ranges. When a diode, IC or transistor is biased on with the higher voltage the current flowing in the semiconductor will upset the reading. Some meters (Fluke is one) have a little diode symbol on the ohm range switch to let you know it will bias on a semiconductor. When the manual asks you measure >100k any value larger than 100k is OK. I would measure this on the 200k range because the lower ranges will just read overrange. If the meter reads overrange on the 200k range I know the reading is over 200k. Sometimes you will have to wait a second or two for any capacitors to charge before the reading is stable.<BR>
<BR>
BTW the ranges that allow the biasing on semiconductors can be used for matching diodes. Remember the meter is measuring the voltage across the test leads so the number is the forward voltage drop of the diode or Vf. Just select diodes that have the same reading.<BR>
<BR>
Don Brown<BR>
KD5NDB</FONT></HTML>
--part1_18b.1486997.29620c9f_boundary--