[Boatanchors] Voltage Calibration with 1000 ohm/volt vs 20000
ohm/volt me...
J. Forster
jfor at quik.com
Mon May 23 22:36:29 EDT 2005
If you have a 20,000 ohm/volt meter and need a 1000 ohm/volt meter, why don't
you just put a resistor in parallel with the meter. For example:
On a 250 Volt scale:
1000 ohm/volt x 250 V = 250 K
20,000 ohm/volt x 250 V = 5,000,000 ohms.
So, what R do you need to parallel a 5,000,000 with to get 250K?
R par = ( R1 x R2 ) / (R1 + R2) ===> so Rpar x (R1 + R2) = R1 x R2
Rpar = 250,000
R1 = 5,000,000
Solve for R2, the resistor you want to put in parallel with the meter.
-John
Flowertime01 at wmconnect.com wrote:
> Hi All
>
> I want to calibrate my TV7-U tester and want to know
> how much of a differance in voltage reading would I
> get using a 1000 ohm/volt mster (called for and dont
> have) vs a simson 260 @ 20000 ohm/volt meter. I have
> seen this a lot with the older receivers also
>
> Mike