Hi Paulo PY3BKT Serafini:
> My Tektronix 2465B + DMM is in trouble
> When turned power on, the TRID D flashes and the CRT is off.
> Pressing A / B TRIGGER several LEDs light but the CRT remains off.
> The voltages + 5V, -5V, -8V, + 10V, + 15V, -15V, -42.4V and + 87V measured
> in J119 are correct.
> Can someone help me?
First, is the CRT filament lit?
This is made in the hV supply, usually.
If if is not lit, that is your problem, the hV section.
If it is lit you need to check the high Voltage if you can, if OK
examine the board that has the "logic" on it.
It may not be "Booting up" or whatever it does when first turned on.
I do not know that 'scope, just how to trouble shoot.
Does the DMM work normally, if not there may still be a low Voltage
power supply problem.
--
Ron KA4INM - Youvan's corollary:
Every action results in unwanted side effects.
More information about the Test-Equipment