On 02/09/2015 12:46 PM, Barry N4BUQ wrote:
> The counter on my generator shows all zeros (both in INTernal mode and EXTernal mode when connected via the front panel BNC jack). I fed the external time-base (rear BNC jack) to a counter and verified 5Mhz is present there.
> My first step will be to verify that a signal is indeed making it to the counter board. If that is present (and I assume it probably is), then I'll have to investigate the counter board(s).
> Is it a valid method to trace out the signal path with a scope? In other words, I think I should be able to verify there are ECL-level and/or TTL-level signals along the path and at least be able to see if there's a point where the signal stops. At that point, then I would start looking for bad components, etc.
> Is that a viable approach? I assume that tracing this out with a scope probe won't give me false readings due to loading, etc. Is that correct?
> Thanks,
In my opinion yes, all you ask about is correct.
--
Ron KA4INM - Youvan's corollary:
Every action results in unwanted side effects.