>I have an HP-410B VTVM.
>
> 1. Can someone tell me why the DC voltage probe
> connection to the HP-410B
> consists of a shielded cable while the Ohms probe
> connection is a just an
> insulated wire - or at least on mine the Ohms probe cable
> is that way (is
> that correct for the Ohms probe?).
>
> So, is the shield really needed on the DC Volts probe?
>
> 2. Also, can someone tell me the original specification
> for the triax cable
> that connects the RF probe (i.e. AC Volts) to the HP-410B?
> It shows an HP
> part number, but I think it was probably a Belden triax
> cable.
>
> I am wondering if any newer more flexible triax cable
> capable of handling
> the voltages work? If so, what should its characteristic
> impedance be (or
> does it matter, as it should be carrying only DC volts
> created from the
> rectified AC/RF)?
>
> Thanks,
> --- Rob Myers, VE3JQL
>
I can't be sure but I think the need for the shield on
the DC lead is due to the presence of a 22 megohm resistor
in the probe body. The purpose is to provide isolation and
reduce capacitive loading on the circuit being measured.
The result is a very high impedance line between the probe
and the meter that I suspect is very vulnerable to induced
voltages. Perhaps someone else has a better explanation.
--
Richard Knoppow
Los Angeles
WB6KBL
dickburk at ix.netcom.com