[Test-Equipment] Problems debugging LCR bridge
Barry
n4buq at knology.net
Mon Apr 23 12:00:48 EDT 2012
> I'm not specifically familiar with this bridge but have
> a couple of general suggestions:
> Is there a half-wave rectifier in it somewhere, for instance
> where does the DC for resistance measurements come from?
> 60hz hum is often from heater to cathode leaks in tubes.
> Substituting tubes is sometimes the only way to find this.
> If the AVC system has a fixed bias source see what that
> looks like.
> Sometimes gassy tubes can cause hum, again the best way
> to find them is substitution.
>
> A good scope will help with this.
>
>
> --
> Richard Knoppow
Hi Richard,
There are two full-wave bridges. One is a 6X4 with a CT transformer and the other is a selenium rectifier.
Yes, it has a fixed bias AGC. I replaced the AGC cap with an electrolytic as I thought that might be part of the problem. I plan to replace it with a 1uF film cap soon. While the AGC line could be the source of the AC, I can't figure why it only occurs when in R mode versus C or L.
As I said, I've disconnected all external inputs to the amplifier so switching from one mode to another should not have any effect (yet it does). If it's a filament to cathode problem, it should show up no matter what the mode.
I thought the selenium rectifier might be inducing something to the EF86 (not likely but I'm grabbing at straws...) so I enabled the selenium rectifuer while in C mode but that still doesn't cause this AC waveform to appear when in C mode.
I also tried shielding the EF86 with no change.
Still can't find it...
Thanks,
Barry - N4BUQ