[Test-Equipment] Question regarding antique LCR bridge
Richard Knoppow
1oldlens1 at ix.netcom.com
Tue Apr 10 16:01:31 EDT 2012
----- Original Message -----
From: "Barry" <n4buq at knology.net>
To: <n4buq at knology.net>; "Discussion of Electronic Test
Equipment" <test-equipment at mailman.qth.net>;
<rbethman at comcast.net>
Sent: Tuesday, April 10, 2012 12:20 PM
Subject: Re: [Test-Equipment] Question regarding antique LCR
bridge
> Not quite. Are you saying that I should be able to
> measure wirewounds more effectively with the bridge than
> other types of resistors? I haven't tried a wirewound but
> will do that and see. Carbon comps were common at that
> time so if it wasn't designed to work with those, wouldn't
> it have stated so in the manual?
>
> Thanks,
> Barry - N4BUQ
I am a little puzzled by this. Most wire wound
resistors have some reactance. A DC measurement does not
show the reactance. It is often useful to know the DC
resistance of reactive components like coils, chokes,
leakage of capacitors, etc. I don't know of any impedance
bridges that do not measure DC resistance. It sounds like
your bridge uses somewhat raw DC, that is DC with a ripple
component, to measure resistance but uses a chopper on the
input of the amplifier so that an AC amplifier can be used
as a null detector. I can't figure out any other reason for
doing this since an AC null detector would work fine for 60
Hz. If no DC is needed why bother with rectifying the AC and
then chopping it back to AC? Chopper stabilized amplifiers
for DC were very common at one time, one can find them in
many DC meters. Its a method of avoiding drift which is
difficult to eliminate from a direct coupled amplifier. Its
also an economical way to provide a high-gain null detector
for DC where an AC null detector is already there.
I don't think there is really much difference between
resistors made now and those of the 1960s other than film
resistors of various sorts have pretty much replaced
composition types. While its commonly thought that
composition resistors are less reactive than film types
measurements show that for low or medium values they are
not. A very old example can be found in the instruction
book for the Boonton 250 RX-Meter, which is intended for
making just such measurements. A sample chart shows the
relative reactance of several types of resistors, the carbon
film one having less reactance than the composition one. I
have confirmed this using my RX meter.
I think your bridge is making conventional DC
resistance measurements.
BTW, while thermoelectric currents do very definitely
affect resistance measurements they are not significant for
values above perhaps an ohm. As the other poster stated
very low resistance (perhaps lower than around ten ohms)
should be measured using a Kelvin Double Bridge or other
four-terminal arrangement. Where thermal currents are of
concern as in extremely accurate comparison with standard
resistors special methods of connection are used to reduce
them or balance them out. There is information on this from
several sources including the old NBS handbook on electrical
measurements. Such precautions are necessary only under very
special conditions and are of no significance when measuring
normal electronic components.
--
Richard Knoppow
Los Angeles
WB6KBL
dickburk at ix.netcom.com