>> If the capacitors were perfect (zero leakage), the two
>> are equal.
>>
>> With real capacitors, option A with a bleeder string to
>> equalize the
>> capacitor voltages would be preferable.
>>
>> Why don't you buy a single cap to do the job and save a
>> bunch of
>> trouble.
>> -John
>>
>
> This is an attempt to replace a "standard capacitor" in a
> General Radio Model 1617A capacitance bridge. The
> original capacitor shorted. I found some 0.5uF, 0.5% caps
> that are rated at 250V so I figured making an array out of
> nine of them would get me 0.5uF at 750V at 0.5% (or
> hopefully better).
>
> I think adding bleeder resistors would upset the balance
> in this application, right?
>
> I have my sights set on a true GR standard capacitor (a
> 1409-X); however, I think at best it's only rated at 500V
> and this bridge can develop 600V of bias so I'm not sure I
> can use that as a replacement. I do plan to use it to
> check the accuracy of my home-made standard.
>
> Thanks!
> Barry - N4BUQ
>
I really think the use of balancing or equalizing
resistors is a non-issue here. These are very low leakage
film caps, mostly equalizing resistors are used with
seriesed electrolytics to prevent excessive voltage across
one _when they charge up_, after that the current is fairly
low. Film caps have such low leakage that I don't think
charging voltage or static voltage can cause any problems at
all.
When putting capacitors in either series of parallel the
errors in their values are averaged, however, that assumes a
normal distribution of error and my not be true of a small
sample, especially if they are all from the same
manufacturing lot. You must really measure them to see at
least in what direction the error is. If you have a known
precise capacitor of the right value, no matter what its
voltage rating, you can measure to a very high degree of
accuracy by the substitution method where the cap becomes
the standard. The overall accuracy depends on the precision
of the measuring instrument, its accuracy is of no
importance since that is established at the desired value by
the comparison standard.
I think in practice there is no significant difference
between the two series-parallel arrangements. As a side
note, it would be interesting to know what is in the
original GR standard cap. The catalogue shows caps up to 500
volts and some to 700 for much lower values of capacitance
than the one in the bridge. It may be that the bridge
standard is a specially made single stack cap or its
possible its made up of two seriesed catalogue standard caps
in a single case. Since the cap is bad you could open the
case to find out what is in it. The fact that its shorted
suggests its a single cap since its unlikely that both caps
of a series would short (but possible).
The accuracy of the bridge with the replacement cap can
be established by measuring a known accurate cap. If the
standard cap is off value the error will be the same on all
ranges using that cap. The _precision_ of the bridge will
not be affected.
--