[Test-Equipment] WTD: Help with old scope problem
J. Forster
jfor at quik.com
Sun Oct 10 15:13:39 EDT 2010
> Hello John,
[snip]
> - Focus has some effecto on the trace but not intensity (strap on rear OK)
> - With 9 VAC applied to the H input on the back, the spot becomes a
line, so, I assume that the H amplifier is OK.
Does the length of the line vary as you adjust the Horiz Gain (Rear Panel)?
> - With 6 VAC applied to the V input, I get a circular pattern (same
> phase) and multiple patterns with a higher frequency.
> - With the voltage calibrator connected to the V input, I get two bars
> with some hint of a square wave. I image due to the low frequency of the
> H input.
This was with the 60 Hz applied to the Ext Horiz or not?
> Both voltage controls (V and calibrator) displace the bars.
> - Neither time/division nor trigger controls have effect on the display.
Try rotating the Time/Div and Trig through it's full range. Do you get any
sweep while rotating the knob?
Set the Horiz Disp to X5 Mag. What is the voltage at pin 8 of V220B with
the Hor Pos set fully CCW and CW?
> - Any way to narrow it down?
Yes, follow the instructions above and report the numbers.
Best,
-John
============