[Test-Equipment] WTD: Help with old scope problem
WA5CAB at cs.com
WA5CAB at cs.com
Fri Oct 8 14:08:15 EDT 2010
Joe,
If you don't get any horizontal movement even from the external inputs then
the problem is between them and the CRT deflection plates. As the vertical
works, the problem must be in the horizontal amp. But I wouldn't know
whether that is one stage or several.
In a message dated 10/8/2010 12:55:48 PM Central Daylight Time,
josepho at shaw.ca writes:
>
> Thank you Bob, the rest of the messages received agree with a problem on
> the
> horizontal sweep, question what is the guilty component?
> Joe
> ----- Original Message -----
> From: <WA5CAB at cs.com>
> To: <test-equipment at mailman.qth.net>
> Sent: Thursday, October 07, 2010 4:05 PM
> Subject: Re: [Test-Equipment] WTD: Help with old scope problem
>
>
> >Joe,
> >
> >I'm not familiar with the 310 but your last point (spot splits in two) is
> >consistant with loss of horizontal sweep. The cal signal is a square
> >wave.
> >The vertical amp is still moving the trace up and down as it should.
> >
> >In a message dated 10/7/2010 5:55:42 PM Central Daylight Time,
> >xyzzypdx at gmail.com writes:
> >>>Hello
> >>>For a number of years Ihave been using a Tek 310A on my bench. I like
> it
> >>on spite of its age and limited BW because of its small size and sharp
> >>trace. Recently,after not have been used for some months, presnted a
> >>number of
> >>problems.
> >>>- There is no horizontal trace, only a spot.
> >>>- The spot can be moved in both directions (V&H).
> >>>- The focus control has only limited effect on the spot.
> >>>- The intensity control has no effect on it.
> >>>- The tubes related to the horizontal and sweep sections check OK.
> >>>- There are no broken, burnt or overheated components.
> >>>- LV voltages are OK but I cannot get a reading on the HV test point
> >>(5642 rectifiers).
> >>>- If I apply a signal from the voltage calibrator, the spot splits on
> >>two.
> >>>Any suggestions as to the cause of these problems (and fixes) will be
> >>greatly appreciated.
> >>>Thanks
> >>>Joe Orgnero VE7LBI
> >>
>
Robert & Susan Downs - Houston
wa5cab dot com (Web Store)
MVPA 9480
More information about the Test-Equipment