[Test-Equipment] WTD: Help with old scope problem
WA5CAB at cs.com
WA5CAB at cs.com
Thu Oct 7 19:05:43 EDT 2010
Joe,
I'm not familiar with the 310 but your last point (spot splits in two) is
consistant with loss of horizontal sweep. The cal signal is a square wave.
The vertical amp is still moving the trace up and down as it should.
In a message dated 10/7/2010 5:55:42 PM Central Daylight Time,
xyzzypdx at gmail.com writes:
> >Hello
> >For a number of years Ihave been using a Tek 310A on my bench. I like it
> on spite of its age and limited BW because of its small size and sharp
> trace. Recently,after not have been used for some months, presnted a number of
> problems.
> >- There is no horizontal trace, only a spot.
> >- The spot can be moved in both directions (V&H).
> >- The focus control has only limited effect on the spot.
> >- The intensity control has no effect on it.
> >- The tubes related to the horizontal and sweep sections check OK.
> >- There are no broken, burnt or overheated components.
> >- LV voltages are OK but I cannot get a reading on the HV test point
> (5642 rectifiers).
> >- If I apply a signal from the voltage calibrator, the spot splits on
> two.
> >Any suggestions as to the cause of these problems (and fixes) will be
> greatly appreciated.
> >Thanks
> >Joe Orgnero VE7LBI
>
Robert Downs - Houston
wa5cab dot com (Web Store)
MVPA 9480
More information about the Test-Equipment
mailing list