[Test-Equipment] HP-3335A no output...

GandalfG8 at aol.com GandalfG8 at aol.com
Thu Oct 15 15:46:08 EDT 2009


 
In a message dated 15/10/2009 14:54:42 GMT Daylight Time, gelabel at gmx.net  
writes:

Now I  have the 21+ MB PDF manual. so I just have to find the time to read 
it
and  to start to fix the generator...


------------------
Gald you downloaded it ok, and good luck:-)
 
I know it's been suggested that you may have a supply fault, and I wouldn't 
 disagree with that possibility, but I still feel the easiest first test  
would be to check the input into the attenuator, especially as you say it  
seems to function ok other than lack of output.
On the 3336 that attenuator is right at the end of the chain, just before  
the output sockets, and I'm pretty sure from the manual that it's the same 
for  the 3335.
Again on the 3336, the input to the attenuator is via a coax plug on a  
flying lead so very easy to disconnect and check for signal on the cable.
It's a very quick check, much quicker to do than write about:-), obviously  
if the signal is missing at that point it's time to start checking the  
supply rails.
 
regards
 
Nigel
GM8PZR