[Test-Equipment] HP-3335A no output...
GandalfG8 at aol.com
GandalfG8 at aol.com
Tue Oct 13 18:02:29 EDT 2009
In a message dated 13/10/2009 22:52:55 GMT Daylight Time, gelabel at gmx.net
writes:
I have an HP-3335A Level Generator which was placed on a shelf, unpowered,
a few years ago.
At that time it worked ok. Today I have powered it on again, apparently
all is ok, the front
panel behaves as expected, but there is no output, nor from the 50 nor
from the 75 ohm
BNC connectors...
Changing the level from the front panel produces some relay clicks.
Any clues on what it can have happened, and what to check ?
------------------
Hi Alberto
I don't know how similar the 3335 is to the 3336 but a big problem on some
of those was one version of the output attenuator where pads on the circuit
board were used to form part of the relay contacts that switched the
attenuator sections.
The first check I would suggest is to open it up and check the signal into
the attenuator. If the signal is there at least you know where the problem
lies, even if fixing it might not be so easy.
I have a pdf ops and service manual if you need it, although it's probably
available as a download online.
regards
Nigel