[Test-Equipment] HP-606A Calibration question
Richard Knoppow
dickburk at ix.netcom.com
Wed May 27 10:57:46 EDT 2009
----- Original Message -----
From: "Barry" <n4buq at knology.net>
To: "Discussion of Electronic Test Equipment"
<test-equipment at mailman.qth.net>
Sent: Wednesday, May 27, 2009 5:24 AM
Subject: Re: [Test-Equipment] HP-606A Calibration question
>
>> No, there is a unique solution, not multiple
>> possibilities. This can
>> be verified
>> by using a resonance calculator such as
>> http://www.daycounter.com/Calculators/LC-Resonance-Calculator.phtml
>> and entering some different values.
>
> I should have realized that. I was thinking there was a
> random variable
> involved in this that would let the L and C "move" and
> stay in resonance,
> but that's not the case. At any given position on the
> dial (and hence the
> variable cap), the other "variables" are fixed so the L
> and C must be
> specific values for resonance. Thanks for straightening
> me out on that one.
> I guess I was just grabbing at straws trying to arrive at
> a reason why this
> thing is so far off calibration at it interior markings.
>
>>
>> John
>> KK6IL
>
> Barry - N4BUQ
This is being made more complicated than necessary.
There only a couple of things that will cause the tuning
"law" to be far off. One is that the dial is in the wrong
place, the other is that the tuning capacitor is damaged in
some way. Examine it carefully to be sure the stators are
parallel with the rotor and evenly spaced. If they are then
adjusting the dial should bring it in. As I described before
there are now three adjustments: 1, the L; 2, the C;, 3, the
dial position. These three must be juggled to bring the dial
in. Set the L and C for the band ends, check the dial at
intervals along the band to find where the error is
greatest. Slip the dial a little one way or the other and do
the L and C again. Then check the center range to see if you
made the error larger or smaller. If smaller then continue
to slip the dial in the same direction, if larger slip it
the other way. Keep slipping it incrementally and then
resetting the band ends. Unless the cap is damaged you
should be able to bring the dial in very closely all over.
When you have done one band clamp the dial firmly. The other
bands should also be in when the L and C are adjusted at the
band ends. This really only needs to be done once.
To those suggesting looking for set screw marks there
are none on a 606A or B. The dial is clamped to the shaft by
a ring with longitudinal screws and there are no marks.
The dials for the 606A/B are not individually
calibrated so even if someone changed the dial it should
still come into specs.
I have no idea how a dial could be moved but its pretty
easy and a lot of people work on this stuff who, bluntly,
don't know what they are doing.
Using a counter to set the frequencies makes the job a
lot simpler than trying to use a crystal calibrator.
If, after doing the calibration procedure carefully the
thing stils won't come in the trouble is likely the tuning
caps. They are hard to see when the generator is assembled.
If you have this problem write and I will describe how to
get at the caps. Its a lot easier than it looks.
--
Richard Knoppow
Los Angeles
WB6KBL
dickburk at ix.netcom.com