 When I first got mine, and not knowing better, I thought that the pegging
meter phenomenon was a sign that the instrument was broken. 

 So, I disconnected the original meter ( 100 uA FSD ) and, in its place,
hooked up an external 0-500 uA meter, allowing for different meter internal
resistances.

 The external meter would go over to about 400 uA on power-up, settle
towards zero after a few seconds, then do it again before finally zeroing.
This seems to be normal.

 Here's my best guess as to what's happening.  There's a big 1200 microfarad
de-coupling cap in the DC negative feedback loop. This is probably necessary
in order to get the frequency response down to 5 Hz. Until it charges up (
thorugh an 8.2K resistor ) and the loop can correct itself, theres an
initial rush of current through the meter movement. I still can't account
for the second 'peg', though.

 It would be interesting to know if anybody has devised a way to prevent the
pegging. Perhaps using some component or trick that wasn't available back
when the instrument was designed?

Ken


This e-mail may contain confidential and/or privileged information for the sole use of the intended recipient. Any review or distribution by anyone other than the person for whom it was originally intended is strictly prohibited. If you have received this e-mail in error, please contact the sender and delete all copies. Opinions, conclusions or other information contained in this e-mail may not be that of the organization.