At 10:03 PM 11/25/2007, DW Holtman wrote:
>I recently acquired a HP-334A. It works in the Voltmeter mode. In
>the distortion mode, it will not null. The balance knobs or main
>frequency dial has no effect on trying to null the unit.
>
>Does anybody have any suggestions on where is a good place to start
>trouble shooting?
If you don't have the manual, download the PDF from the Agilent site:
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=811981&lc=eng&cc=US&nfr=-11143.0.00
See the troubleshooting table on page 5-13, symptom C.
On the A6 bridge balance board are five lamps (DS1 - DS5) coupled to
five photocells (V1 - V5). I would think DS1 might be bad.