[Test-Equipment] HP-608D parts for sale
Richard Knoppow
dickburk at ix.netcom.com
Wed Dec 5 21:45:51 EST 2007
----- Original Message -----
From: "wabate" <wabate at verizon.net>
To: "Test Equipment" <test-equipment at mailman.qth.net>
Sent: Tuesday, December 04, 2007 8:57 AM
Subject: [Test-Equipment] HP-608D parts for sale
> My trusty 608-D finally died yesterday when the output
> attenuator failed. In reading the manual it does not seem
> practical to repair nor do I have the equipment necessary
> to recalibrate the output. So I'm offering it out to the
> list as parts. This is a nice clean generator and
> everything works except the output attenuator. Meters are
> not cracked, knobs are all there, tubes all appear to be
> fine, cabinet is pretty nice, nice manual, crystal, etc.
>
> Let me know what you need. Too heavy to ship. I'm near
> Philly.
>
> Bill, K3PGB
>
Its been more than 40 years since I worked on 608s
for -hp- but I still remember changing the resistors in the
probes. The originals were tiny deposited carbon but modern
metal film resistors should work fine. Instructions are in
the handbooks available at BAMA. If you make sure to get
everything back in the same place the output calibration
should not change much.
The main cause of probe problems was burning them out
by connecting the generator to the output of a transmitter
for some reason. Its also possible for the capacitor to
short. My memory is that this was not very common.
A couple of notes from memory:
V-5, the 6CL6 output regulator, is quite sensitive to
heater to cathode leakage. We had to select tubes that would
maintain constant output level with varying line voltage. I
was never able to find a cure for this although my memory
still itches that I saw some method of compensating for this
leakage somewhere.
Note 2: The oscillator uses conventional L-C tanks. The
inductors are held in place on the forms by coil dope. This
often cracked and loosened even in generators which were
only a few years old causing the frequency to be off. The
curious thing is that no one at the service center seemed to
realize that the coils were adjustable and the routine was
to cut new dials. Not necessary at all. Simply adjust the
inductors at the low end and the trimmer caps at the high
end of each band. some juggling is necessary. The lower
frequency coils have shorted turn adjusters on the coil
form. They can break loose and move. Also, the coils
themselves break free of the form. Squeeze them into
calibration if they are loose and fasten them with coil
dope. If the coils are tight the error will probably not be
large and can be corrected by moving the shorted turn a bit.
Again, once its set lock it with coil dope. The higher
frequency coils must be adjusted by adjusting the turn
separation by squeezing or stretching a bit. Again, once in
place dope them down. The calibration must be checked with
the side cover on. Its a PITA to have to remove it to make a
small adjustment and then put it back but, once done, it
won't need attention again for a long time. After getting
the thing adjusted close it up and let it run for at least
48 hours and re-check the calibration. You may have to make
a small adjustment at the top end of each band. When
properly calibrated the dial tracking is excellent and
should be right on the mark everywhere.
Defective oscillator tubes can cause calibration errors
so another tube should be tried if the calibration was right
and is serously off with another tube. I had to do this
occasionally even though I was using -hp- inspected tubes.
The rule with all older (vacuum tube) -hp- gear is that
anything that was not better than half-specs was broken.
---
Richard Knoppow
Los Angeles, CA, USA
dickburk at ix.netcom.com
More information about the Test-Equipment
mailing list