I have been following the thread on old test equipment, which changed to a
discussion on standard cells.
Here is my question:
I have a Guildline 9152/4 Standard Cell enclosure that is fitted with 4
saturated standard cells. The cells are held at a temperature of 30 +/- 0.01
C.
At one time this type of cell was used as a transfer standard, before being
replaced by solid state units like the Fluke 732.
When I measure the cells I find that two of pairs of the cells agree with
each other by 1 uV. But the pairs disagree with each other by 25 uV.
I do not have the history on this unit. Is this a normal variation for this
type of standard cell ?
The voltages are:
Cell 1 1.018129
Cell 2 1.018130
Cell 3 1.018154
Cell 4 1.018155
Measured with a Fluke 8506 or HP 3456
Can anybody help ??
Regards,
John Barnes