[Test-Equipment] Tracking generator for HP 8569

Dave Emery [email protected]
Fri, 20 Feb 2004 16:02:15 -0500


On Fri, Feb 20, 2004 at 02:21:53PM +0100, AdB wrote:
> Hi all,
> 
>   does it exist a matched tracking generator for the Spectrum Analyzer
> HP-8569  ?
> 
> I have found references to the HP-8444A with opt.059, but its max frequency
> is 1.5 GHz, far less than the 18 GHz of the 8569.


	I don't know of very many tracking generators for spectrum
analyzers that go much over 3 ghz.

	They are generally most useful in the HF, VHF and UHF bands
(and indeed VERY useful there, for network analysis type measurements,
like VSWR, loss and gain, and so forth - nearly indispensible for
tuning filters, diplexers, antennas and so forth).

	But higher up in frequency it gets much more expensive to
generate signals at the required power levels and issues such as cable
loss become more and more significant.    Most tracking generators have
a tap (directional coupler) and detector at the output connector to
level the output - which means it is flat at the calibrated level
at the connector.   This is fine assuming one's cables are reasonably
low loss, but falls apart when microwave frequencies are involved
and losses can be upwards of 10 db and not flat with frequency.

	Many sweepers have provision for external detectors to level
the output... but such complexity is more difficult to include in
a tightly packaged spectrum analyzer.

> 
> Thanks in advance
> 
> Alberto  I2PHD
> 
> 
> _______________________________________________
> Test-Equipment mailing list
> [email protected]
> http://mailman.qth.net/mailman/listinfo/test-equipment

-- 
   Dave Emery N1PRE,  [email protected]  DIE Consulting, Weston, Mass 02493