[Test-Equipment] HP141T+ HP8554L+ HP8552B Question
Scott Townley
[email protected]
Fri, 20 Sep 2002 05:41:37 -0700
Noise figure is roughly 24dB on the 141T series.
Noise floor depends on the resolution bandwidth. Approx. FkTB*1.1 where F
is the noise figure and B is the rbw, T=290K and k=Boltzmann's constant
1.38E-23
And that also answers the sensitivity question
At 12:20 2002-09-20 +0100, John Mortlock wrote:
>I have an opportunity to buy a spectrum analyzer set-up consisting of
>HP141T+ HP8554L+ HP8552B.
>
>Can someone answer the following for me please:
>
>1 What is the maximum sensitivity of this set-up (what is the smallest
>measurable signal in dBm) ?
>
>2. What is the noise floor ?
>
>
>Thank you.
>
>