Sounds like a regulator problem. As it heats up the supply voltage is
dropping. Have you checked to see that the 10Mhz Ref Out amplitude
doesn't change? That would eliminate the oven/osc.
AFAICR the gen has an autoleveler cct. One the bypass caps has probably
gone leaky , dragging down the reference voltage. Check the amplitude
pot wiper to see that the voltage is unchanging.
Give my your fax number and I'll see what I can do (after I find the manual).
Robert
[email protected] wrote:
>
> Hi guys -
>
> I've been fighting with a problem on my IFR AM/FM-1000A service monitor and
> it turns out that my service manual only has about half the schematics in the
> binder -grrrrrrr! It looks like I'll need to find copies of the schematics
> for the "AM/FM Generator" module and the "10 Mhz Oven/Osc" module.
>
> IFR has indicated they are no longer "supporting" this unit - even to the
> point of not being willing to make courtesy photocopies for me (strange).
> Oh, well... does anyone have a complete service manual that has those pages?
>
> BTW - the problem I'm seeing is a smooth drop off in generator output as much
> as 20 dB shortly after the unit is turned on. Ring a bell for anyone?
>
> Steve WD8DAS
>
> [email protected]
>