Very true, I just had to discover this for myself a couple of weeks ago.
Had a 10811A that would oscillate but couldn't be brought into calibration.
I replaced the 109-degree (if I remember correctly) fuse with a similar NTE
replacement part. The replacement was slightly larger, so I had to put a
piece of spaghetti over one lead to avoid shorts. With these thermal fuses,
you don't want to bend the leads sharply near the body during installation.
Coincidentally, I later found a "hint & kink" in a recent QST covering this
very issue.
I don't see how that fuse could really protect anything, anyway... it's a
long way from the inner oven. Replacing it with a straight piece of wire
sounds reasonable.
-- jm
> -----Original Message-----
> From: [email protected]
> [mailto:[email protected]]On Behalf Of Mike Feher
> Sent: Sunday, December 15, 2002 6:23 AM
> To: Bob Parnass, AJ9S; [email protected]
> Subject: Re: [Test-Equipment] HP 10811A - 5335A
>
>
>
> Talking about the 10811A series and its predecessors, I have
> found that the
> thermal fuse in these often goes bad (opens up).
>