[R-390] Official specs

Jim Whartenby old_radio at aol.com
Tue Oct 22 18:36:21 EDT 2024


Larry
I built a test fixture that is essentially two DA-121's connected back to back.  Photos and drawing are enclosed.  This does the conversion from 50 ohms to 125 ohms and then back to 50 ohms.  I used 1% resistors to make the attenuator circuit with the values close to those found here:  https://k7mem.com/Res_Attenuator.html 

The closest I could come to the 64.18 ohms result from the attenuator calculator was 63.9 ohms.  This is from the parallel combination of 3 each 237 ohm in parallel with a 1k, in parallel with a 499 ohm resistor.  Five resistors in parallel, all 1% resistors.  The result was 63.85 ohms, a 0.5% error.  The sub for the 96.83 ohm resistor is a 100 ohm 1% resistor (3% error) and the sub for the R-390's 125 ohm impedance was a 121 ohm 1% resistor (3% error).  This is still much better then the 5% resistors used in the original DA-121.

For a test oscillator I used a Helper SM-1000 signal generator and measured the insertion loss with a Stoddart NM-25T frequency selective voltmeter.  The insertion loss was measured at 10 MHz using two 4 foot BNC RG-58 coax cables from Pomona Electric.  4 foot of coax from the SM-1000 to the test fixture and another 4 feet from the test fixture to the NM-25T.

The SG was set for a reading of 30 dB on the NM-25T signal strength meter when measuring a BNC through connection and then measured 11 dB when the test fixture was installed in place of the BNC through.   The insertion loss for the test fixture is 19 dB.  Dividing this by two since there are essentially two DA-121s back to back gives an insertion loss of about 9.5 dB for a single DA-121.  This closely agrees with the attenuator calculator findings.  

So it seems that your DC circuit calculation do not agree with the RF measurements.  Transmission lines behave differently then DC circuits.  You calculate a 5 dB insertion loss, I measure a 9.5 dB insertion loss.

Here is an experiment that you can try.  Insert a 50 ohm resistor in parallel with the 50 ohm coax.  What do you think will happen?  Perhaps nothing since the coax is 50 ohms and the resistor is also 50 ohms?  In reality, the coax has reactive elements, parallel capacitance and series inductance that make up the coax impedance.  Neither of which will dissipate the signal carried on the coax.  The only losses are from the resistance of the conductors that make up the coax.  Adding a parallel resistor will attenuate the signal to the receiver by 3 dB.

If anyone on this list wants to make their own version of the DA-121, I can supply the resistor values I used for a token $2 plus postage.  Just DM me with your address and if you want one or two resistor sets.

Regards, Jim
Logic: Method used to arrive at the wrong conclusion, with confidence.  Murphy 

    On Friday, October 18, 2024 at 05:36:08 AM CDT, Larry Haney <larry41gm2 at gmail.com> wrote:   

 Hi Jim,  I just checked and I only have 1 da-121.  As for insertion loss, my coax is very short and the connections are very good so the loss there would not be possible for me to measure.  Now for the insertion loss due to impedance mismatch (due to resistance variations) would also not be possible for me to measure, as I don't have the equipment required for that.  But, because the 3 resistors in the circuit are very close to the required values for a perfect 50 ohm match to the sig gen, I am sure that the insertion loss due to that very slight  impedance mismatch is extremely small.  I have no way to measure that loss as I don't have the 3 exact value resistors to compare it to.  I could calculate it, but I believe that would be a waste of time without being able to measure it.
After all the input you have given me and the research just done, I'm satisfied with my current measurements and calculations (IE: the output voltage of the da-121 is 56% of the input voltage when the load is 125 ohms).
My biggest concern about making snr measurements is for those folks that don't have a recently calibrated sig gen or calibrated rms AC voltmeter to verify their readings with.
Regards, Larry
On Thu, Oct 17, 2024 at 1:55 PM Jim Whartenby <old_radio at aol.com> wrote:

LarryNo, just one SG and one 125 ohm load.  You should be able to determine the total loss through two DA-121 attenuators connected back to back with an o'scope and then divide the loss by two to solve for the insertion loss.JimLogic: Method used to arrive at the wrong conclusion, with confidence.  Murphy
  
-------------- next part --------------
A non-text attachment was scrubbed...
Name: DA-121 bottom.jfif
Type: image/jpeg
Size: 113851 bytes
Desc: not available
URL: <http://mailman.qth.net/pipermail/r-390/attachments/20241022/2132909f/attachment.jfif>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: DA-121 top.jfif
Type: image/jpeg
Size: 123165 bytes
Desc: not available
URL: <http://mailman.qth.net/pipermail/r-390/attachments/20241022/2132909f/attachment-0001.jfif>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: DA-121.pdf
Type: application/pdf
Size: 53744 bytes
Desc: not available
URL: <http://mailman.qth.net/pipermail/r-390/attachments/20241022/2132909f/attachment.pdf>


More information about the R-390 mailing list