Renee,

 

Thanks for your thoughts.  What do you use to heat the C551 can to disassemble it?

 

Thanks again.

Gary

 

From: Renee K6FSB <k6fsb.1 at gmail.com> 
Sent: March 6, 2022 10:16 AM
To: Jordan Arndt <Outposter30 at shaw.ca>; gary.biasini at shaw.ca; R390A List <R-390 at mailman.qth.net>
Cc: 'R-390 Forum' <R-390 at mailman.qth.net>
Subject: Re: [R-390] Carrier Level Meter Issues

 

Hi All-
I replace C551 on sight....on 16 (IIRC) IF decks I have never found one that does not have leakage current ( maybe just my bad luck, tested on Sencor LC102)...oil/paper....i went with mylar anywhere between 1.5uf or 1.8 uf ( I liked these best) to 2.2uf.
I machined small rectangular aluminum cans (with teflon feed thru for the leads) for them to fit in the original space. and as Jordan suggests the 1.5uf will fit in the original can......I drill a small hole before heating to disassemble the can. I do not like hot oil everywhere......
some of the postage stamp Micas are starting to have shorting issues due to impurities, moisture intrusion and what is called silver migration...so do not rule them out because they are mica.
Renée, K6FSB
 

On 3/6/22 4:25 AM, Jordan Arndt wrote:

Hi again group... 

Any leakage of C551 will dramatically alter the AGC characteristics, with the results of leakage being seen as a longer to much longer AGC recovery time and a lower and "softer" AGC voltage. 

I rebuilt C551 in my three 390As using non-polarized 1.5uF @ 450Vdc film caps which just fit inside the original square can. The result is a "stiffer" AGC voltage, and a noticeably quicker return to receive than with a 2uF cap. 

I use a restored Sprague TO-6 to detect leakage, and these film caps show no measurable leakage at all, where all of the original C551 caps clearly showed varying degrees of leakage. I'd expect that most, if not all, of the original C551 caps are leaky by now and should be replaced/rebuilt... 

73...Jordan VE6ZT 

----- Original Message ----- From:  <mailto:gary.biasini at shaw.ca> <gary.biasini at shaw.ca> 
To: "'Larry H'"  <mailto:larry41gm at gmail.com> <larry41gm at gmail.com> 
Cc: "'R-390 Forum'"  <mailto:R-390 at mailman.qth.net> <R-390 at mailman.qth.net> 
Sent: Saturday, March 05, 2022 10:13 PM 
Subject: Re: [R-390] Carrier Level Meter Issues 





Larry, 



Thanks. In fact, I started on one tonight which will be the subject of a separate post once I do a bit more work. 



I checked the manual for the IT-11 and in fact, when properly calibrated, on the setting for “Paper, Mica, etc.” capacitors – i.e. anything other than Electrolytic or Mini-Electrolytic capacitors, the eye will not open if the leakage is more than 2 ua so, the C551 that I tested leaks less than 2ua. 



By the way, the threshold for Mini-‘lytics is 15ua and the threshold for Electrolytics is the 2ma. 



Thanks again for your response. 


______________________________________________________________ 
R-390 mailing list 
Home: http://mailman.qth.net/mailman/listinfo/r-390 
Help: http://mailman.qth.net/mmfaq.htm 
Post: mailto:R-390 at mailman.qth.net 

This list hosted by: http://www.qsl.net 
Please help support this email list: http://www.qsl.net/donate.html

 



More information about the R-390 mailing list