[R-390] Carrier Level Meter Issues
Jordan Arndt
Outposter30 at shaw.ca
Wed Mar 2 15:43:49 EST 2022
Gary...
What are the AGC and the Diode Load voltages when you run that test in Fast
AGC...? Ignore Slow and Med for now...
It sounds like C551 is leaky, the meter should read nearly the same in all
three AGC positions.
73...Jordan VE6ZT
----- Original Message -----
From: <gary.biasini at shaw.ca>
To: "Jacques Fortin" <jacques.f at videotron.ca>; <hamfish at comcast.net>;
<R-390 at mailman.qth.net>
Sent: Wednesday, March 02, 2022 12:38 PM
Subject: Re: [R-390] Carrier Level Meter Issues
Jacques,
Thanks for weighing in. AGC voltage is -5VDC with Bandwidth 16kc, RF gain
fully clockwise, Limiter off.
DB meter starts at 45 in Slow AGC, pins the needle to the right for a second
or so in medium then settles on 58 and does not move in fast AGC. Going back
to medium, no change, back to slow pins the meter for a second and then it
settles at 45.
Regards,
Gary I. Biasini