That 10M is at DC, but there's significant RF at the test point, and the stray capacitance of the usual kind of DMM probe will short it right out.  Heed Larry's suggestion to splice a 100K resistor into the probe tip to decouple the probe from the test point.  This is a useful trick anytime you're looking for DC amid RF.  The resistor will reduce the reading slightly, but you can fudge it out with the voltage divider equations.

These test points show DC because the mixer grid and cathode act as diode detectors.  (Acting in shunt with the signal.)  What you read on the meter is the mathematical average value of the half-wave-rectified RF.  Obviously if you short out the RF there won't be much DC left.