You may be barking up the wrong tree so far as AGC problems are concerned.

 

The most likely failure mode for the AGC bus is a leaky bypass capacitor,
not the AGC delay capacitor C551, which is a very solid unit.

 

Open the jumper between terminals 3 and 4 of TB-102.  Measure the resistance
from Terminal 4 to ground. It should be infinity. If it isn't your problem
is one of the .005 uF bypass capacitors on the AGC bus.

 

If the resistance from Terminal 4 to ground shows open, check C543, C547 and
C548, all of which are "black beauties" and should be replaced in any case.

 