J Adjust the IF gain R519
___ 01 Set the Line Meter OFF
___ 02 Set the Line gain to 0
___ 03 Set the ANT Trim to 0
___ 04 Set the AGC to MED
___ 05 Set the LIMITER to OFF
___ 06 Set the Band Width to 2KHz
___ 07 Set the BFO Pitch to 0
___ 08 Set the Audio Response to WIDE
___ 09 Set the Breakin OFF
___ 10 Set the Function to MGC
___ 11 Set the BFO to OFF
___ 12 Release the Zero Adjust
___ 13 Release the Dial Lock
___ 14 Set the Local Audio to max
___ 15 Set the RF to max
___ 16 Remove P114 from J514
___ 17 Remove P213 from J513
___ 18 Remove P218 from J518
___ 19 Remove J116 adapter from the back panel if necessary
___ 20 Couple P114 to J513
___ 21 Couple P116 to J116
___ 22 Couple J116 to the signal generator RF output
___ 22 Adjust signal generator for 455 KHz output frequency
___ 24 Adjust signal generator for 150 micro volt RF output
___ 25 Adjust signal generator for 30 % audio tone modulation (400 Hz)
___ 26 Meter diode load output for -7 volts DC
___ 27 Place a 600 ohm load across the local audio output
___ 28 Place a 600 ohm load across the line audio output
___ 29 Meter local audio output for 450 milliwatts, 27 db, or 17.3 Volts AC
___ 30 Adjust the IF gain R519 for -7 V DC on the diode load
___ 31 Observe the local audio output level is greater than 400 milliwatts
___ 32 Local Audio should be 17.3 Volts AC across 600 Ohms 450 mw
___ 33 Line Audio should be 2.45 Volts AC across 600 Ohms 10 mw
___ 34 Phone Audio should be .78 Volts AC across 600 Ohms 1 mw
___ 35 Line Audio at .78 Volts across 600 should be Line Meter Zero VU
___ 36 Set the Line Meter to +10
___ 37 Set the signal generator modulation on
___ 38 The Line Meter should read above 0 VU (10 mw)
___ 39 Set the Line Gain off max until the Line Meter reads 0 VU (10 mw)
___ 40 Set the signal generator modulation off
___ 41 Set the Meter Switch to -10
___ 42 Observe 30 db change (20 db on switch plus 10 db on meter scale)
___ 43 The Line Meter should read less than -10 VU (SN + N > 30 DB)
___ 44 Set the Line Meter to OFF
___ 45 Set the Line Gain to 0