Indeed, that ext. counter input is spec'd from 100 mVrms to 1.3Vrms. That
last one is very clearly marked as a "may-not-exceed" in the service manual.
Also, its Zin is 50 ohms, another way to get things messed up.
It leaves the cap on or it gets hosed!
regards,
Mahlon - K4OQ
> ------------------------------------------------------------------------
>
> Subject:
> Re: [R-390] HP 8640B Question
> From:
> 2002tii <bmw2002tii at nerdshack.com>
> Date:
> Sat, 20 Feb 2010 16:07:59 -0500
> To:
> r-390 at mailman.qth.net
>
> To:
> r-390 at mailman.qth.net
>
>
> Bob wrote:
>
> <snip>
>
> That said, most 8640s I've seen that read all zeros don't do so because
> the counter input is set to EXT, but rather because someone once used
> them to count external signals and fried the counter. That is the
> single most common 8640 problem I've seen. So much so that I advise
> folks NEVER to use the 8640's external counter. It comes with a BNC cap
> -- leave it there. Note that the military Option 323 units do not have
> an external counter input, for this very reason.
>
> Best regards,
>
> Don
>