[R-390] R-390A dB meter as as "S" meter
2002tii
bmw2002tii at nerdshack.com
Fri Nov 27 16:18:28 EST 2009
David wrote:
>How do you calibrate your S meter. I was trying to that on a 75S-3 and
>gave up other than to just set the IF gain to give a S9 reading when the
>meter on my HP-606B read 50 uV.
>
>Since the 606B meter reading are accurate only when working into a 50
>ohm load, what load is the input circuit of the receiver? Is it the
>same for all bands? If you use a 50 ohm terminator for the signal
>generator, do you put it at the signal generator end of the coax
>connector or at the receiver end. What is the impedance at the end of
>the piece of coax that connects the receiver to the signal generator?
No, not all receivers have 50 ohm inputs, although there is a modern
trend toward 50 ohms. One would typically get this information from
the manufacturer's spec sheet, but it can be measured with an RF
impedance bridge if you don't have the manual or if you have reason
to believe the manufacturer's figure is not accurate.
Since the "S" scale is a power scale, one would ideally use a
transformer to match the generator to the actual rx input
impedance. -73 dBm (50.2 uV) at the meter would then be S9 for all
radios, regardless of the voltage at the transformer secondary. In
actual practice, it is more common to build a resistive pad to
transform the impedance. In this case, one must figure the insertion
loss into the determination. In either case, one would use 50 ohm
cable from the generator, and put the transformer or pad right at the
radio antenna input terminals.
The balanced input of the R-390A is specified as "125 ohm terminating
impedance." The impedance of the unbalanced input is not
specified. The specified procedure for measuring sensitivity
(NAVSHIPS 0967-063-2010 pp. 4-2 and 4-3) involves connecting an
AN/URM-25( ) generator to the unbalanced input and adjusting the
generator output level for 10 dB quieting of a 30% modulated signal
at 12 specified receive frequencies. The generator level should be
less than 3 uV for 10 dB quieting. (Per Table 1-1, the sensitivity
spec is 5 uV for 10 dB quieting on AM signals; which input is not specified.)
According to the NAVSHIPS manual ((table 5-5, step 17), the R-390A
carrier level meter should read at least "40" with 10 uV from an
AN/URM-25( ) connected to the unbalanced input, and "60" +/- 2 with
100 uV, measured at 15 MHz.
Best regards,
Don
More information about the R-390
mailing list