[R-390] A R-390A carrier level defective -Happy New Year for all !
flowertime01 at wmconnect.com
flowertime01 at wmconnect.com
Sat Dec 26 13:12:01 EST 2009
Sam Rocha, PY1 DKW, Brazil,
The next step is to haul out the basic meter testing procedures (copy off
the web from some place) and test the meter.
These meters do go bad. Mostly a bit of stuff gets wedged in the meter
movement and causes the meter to hang up. You can clean the meters up and save
them. Next problem is stuff sets up in the meter jewels. This to can be
cleaned up. This work is out there at the watch maker level. The meters do come
apart and can be cleaned.
Do test the meter to determine if the problem is internal or external.
If internal, you may just want to let the meter set as is until you get
every thing else done.
You may want to clean and inspect all the subassembly plugs and pins
between the IF deck and the meter. A dirty pin or plug can cause losses and still
allow you to zero the meter.
Roger AI4NI