GM Pete and all,
Point well taken. After spending time considering what was done to T
comp these oscillators, simply removing one of those delicate caps
does seem a bit provocative. I don't know what temp specs were
required of the PTO but the needed temp range had to be greater than
the ~20C over which you tested. A 10 Hz shift is certainly minimal
and would be especially so if linearity was not hurt and the end
points over temp did not significantly shift relative to each
other. Operationally, the change over temp would then be just a
small offset of dial calibration over temp; easily correctable by the
calibrator. Perhaps swap the NPO for the N750 and run it over temp
again to see how it affects T comp. Bottom line - looks like you
have a viable spare.
If you would like better temp performance, observe the necessary
change in C and then come up with a new combo of T comp caps to
maintain the expected original change in C over temp. This assumes
that although the C value of the 370 pF stud mount cap may have
changed, it also assumes that the temp coefficient of that cap did
not change. At this point though we are rather deep in an area of
diminishing returns...
Another thought about T comp needs is that it's my understanding that
many of these receivers were used for RTTY and temp performance may
have been driven totally by those requirements. Basic PTO temp
performance AND the oven may have been needed for long term "set and
forget" tuning stability.
Best regards,
Jon Schlegel
WA3MVM
At 04:52 PM 5/20/2008 +1000, you wrote:
>List.. g'day.
>Calibration /adjustment of the COSMOS pto has been covered fairly
>extensively and removal of one turn from the endpoint coil appears in most
>cases to allow correct 10 turn spread.
>However , having decided to check out one of my spares , I found the
>situation where one turn off the coil and adjustment of the slug did not
>bring the end points within spec--- about 5 kHz low.
>Reasoning that one or more of the capacitor components had gone high, I
>snipped out one of the miniature 10 pf across the whole inductance .
>Result is that the PTO endpoints, after processing, now accurately set and
>linearity spot on.
>No doubt the purists will take me to task on this as the 2 miniature caps in
>parallel: one is N750 and the other NPO.... I could not identify which one
>I removed . The posibility too ,is that the large screw mounted 370 pf has
>changed value cannot be discounted, but the effort on checking this and
>even getting a replacement made the previous option a more viable
>proposition.
>Stability before and after appears not to have changed, and heat
>treatment from room temp. at 50 F to 90F only changed the freq. by a max of
>10 Hz.
>The testing was done with a maximum of convenience...on the bench using a 2
>voltage regulated variable supply giving the 150 and 250V required and the
>output loading to a frequency counter.
>
>Any comment appreciated together with caveats on the empirical procedures.
>No doubt there is the possibility that if unstable the cap or whatever
>might revert to normality or worse--- but them's the breaks !
>
>Cheers
>Pete
>
>
>
>
> Pete D.Williams
>jupete at bigpond.net.au
>METUNG 3904 AUSTRALIA
>
>_____________________________________________________________
>R-390 mailing list
>Home: http://mailman.qth.net/mailman/listinfo/r-390
>Help: http://mailman.qth.net/faq.htm
>Post: mailto:R-390 at mailman.qth.net
>Unsubscribe: http://mailman.qth.net/mailman/options/r-390
More information about the R-390
mailing list