[R-390] I probably shouldn't admit this but...
n4buq at knology.net
n4buq at knology.net
Thu Mar 29 09:04:38 EST 2007
About a year or so ago, I noticed the probe for my HP410B stopped heating
and I was getting no AC readings. After checking, I discovered the probe
has an EA53 (a 6.3VAC filament diode) instead of a 2-01C (a 5VAC filament
diode) and the heater voltage had been set to 5VAC. I increased the
filament voltage to 6.3VAC and the probed started heating again; however,
it appeared that my AC voltage readings were off considerably.
Figuring the EA53 was failing, I shelved it waiting to find either a 2-01C
probe or an affordable EA53. I recently noticed an auction for a probe set
for the 410B but it appeared to have the AC probe from an ME26D/U. I
emailed the owner and asked if he happened to have any EA53s and he did;
however, he said it sounded like something else was wrong with the meter
instead of the diode.
I started checking around on it again. I opened the probe top and fed the
AC input via a regular silicon diode. At 60cps, the results were exactly
the same as the EA53. With the probe still open and the diode powered, I
checked whether the EA53 was working and I confirmed with my scope that it
was functioning as a half-wave rectifier. It appeared something further
upstream from the probe might be wrong.
I then happened to think of something. My test voltages were coming from a
variac which was being fed with an isolation transformer. Since I know
that when measuring line voltages it is not proper to connect the ground
but only use the probe tip, I was using that method to get sample voltages
form the variac. It then occured to me that I was checking the voltages
across an isolated circuit. I connected the ground clip and the meter
again start reading correctly (or nearly so). After running through the
calibration steps, the meter now is quite accurate on AC.
Somethimes I feel like the scarecrow from the Wizard of Oz...