[R-390] OT - Lambda power supply repair
Paul H. Anderson
paul at pdq.com
Tue May 16 11:42:55 EDT 2006
For whatever reason, I like the mid-80's lambda lab power supplies with
the LED volt/amp front meter. I had over the years picked up a few from
various places, some NOS, some beat up. I use these for powering various
projects, including R-392s, R-391 autotune, and my T-195.
A few years ago, I wrote to this list asking about ideas for repair, and
really got no useful feedback. I assumed for a long while that repair was
not feasible due to internal complexity of the switching design.
Here is an example of the type I'm referring to:
<http://login.pdq.com/boatanchors/lambda/lambda_lq-204.jpg>
I recently had some time to spare, so started looking into several common
problems they have, such as:
unlit LED display
wildly flashing voltage/amp levels
osciollating voltage output
and so on, not to mention one with a variety of dead shorts. I feared
fried major components or ICs.
It turns out that besides the primary and secondary electrolytics failing
short, that the other main problem is electrolytic axial caps failing in
large numbers.
I've started rebuilding several units like this, and am finding it to be
easier simply to replace all the caps with equal or slightly higher values
(and higher voltage ratings) from Mouser, et al. Only after doing that is
it worth doing any real in depth testing or repair.
So far, I'm 2 for 3, with #3 being a 45 amp 28V unit that I haven't yet
replaced the axials in (I plan to do this shortly).
In the two unit I tested and repaired fully, more than half of the axial
electrolytic caps tested bad on my ESR meter or ohmmeter, and replacement
of all circuit board mounted electrolytic caps (aside from the filtering
caps) returned it to normal operation. My approach ended up being
wholesale replacement of axials on the circuit boards, and selective
replacement of the large power filtering caps, which are much more
expensive.
My suspicion is that wholesale replacement of electrolytic caps in this
type of units will tend to return them to as-new condition, or at least to
ham hobbyist bench usable condition.
I'll try and get some useful inside pictures of original and replacement
caps for folks to look at. I think most are failing due to long term heat
exposure.
I have not had these types of problems with the modular brick style
lambdas - I don't know why - possibly better thermal controls or a
different style of usage for those.
Anyway, definitely worth a shot if you've got one and been wondering what
on earth is wrong with it.
Please remember to be extremely careful with these when opened. The
primary side is over 200V, and should be treated with extreme caution.
Even when powered off, either normal or failing power supplies may have
high voltages present.
Paul
More information about the R-390
mailing list