You may be deceiving yourself with respect to the sensitivity measurement.
You make no mention of terminating the 8640B signal generator.
To get an accurate reading, the output attenuator of the 8640B (and any
other signal generator) has to be terminated with a specific load resistance --
in this case, 50 ohms. If it is terminated with a high impedance -- i.e., the
unbalanced input of the R-390A -- less current is going to be pulled through
the series element of the attenuator, and the "multiply by" marked on the
skirt of the output attenuator dial isn't going to be correct. What appears to
be 0.2 or 0.3 uV is actually going to be higher.
The BALANCED input is nominally rated at 125 ohms (though this is probably
far from constant either from one band to another or across a signle band). A
50-ohm signal generator connected to the balanced input should therefore
terminated by an 82 ohm resistor (with short leads) connected across the antenna
input terminals.
The impedance of the UNBALANCED input is not indicated in any of the
documentation I have, but it is supposed to be "high." I would therefore terminate
the signal generator with a 51 ohm resistor so that it sees a proper load.
Only when the signal generator is properly terminated do the output signal
readings shown by the panel meter and the switch attenuator correspond with
reality.
Miles, K2CBY