[R-390] DMM bounce
Bob Camp
ham at cq.nu
Mon Jul 12 23:28:43 EDT 2004
Hi
If you are using a trusty VACCUM TUBE instrument then you will have no
problem with vacuum tube impedances.
A modern super duper high speed auto ranging DVM will to just a little
nuts trying to measure the resistance of your 360 Henry inductor. Some
of them at least drive the device under test with a constant current
source. This is fine for a capacitor or resistor. On an inductor the
current ramp gives you a voltage ramp. Up goes the voltage and it auto
ranges. Even if it runs a constant voltage it will still auto range if
it's fast enough. The coil will look like an open circuit at first so
the things little brain decides to go to a higher resistance scale. In
either case you can get a situation where it just bounces back and
forth with no tendency to settle at all.
Simple solution - hook a low voltage power supply (like a D cell) in
series with the DVM and the secondary. That should give you about a
milli amp though the secondary. Unless this is a very low level
transformer the current shouldn't bother it to much.
I use a similar setup for checking caps on the R390. A 60 volt power
supply, a 5 micro amp DC meter and a 22 meg resistor all in series
with the capacitor makes for a pretty sensitive ohm meter / cap leakage
tester.
You can use the same resistor in series trick to keep the current in
the secondary low. In that case I probably would measure the voltage on
the secondary and figure the resistance out by formula. Figure that
kind of stuff out for a living and they make you a Lord ....
Don't know if that helps or not.
Take Care
Bob Camp
KB8TQ
On Jul 12, 2004, at 7:03 PM, Dan Merz wrote:
> Hi, here's a general question related to making resistance
> measurements
> when an inductance is involved, e.g. the winding on an audio
> transformer. Generally, I make such measurements either with my BK
> DMM on a fixed scale setting or with a Simpson 260 old-style
> multimeter, and I usually have no problem. But recently I repaired an
> audio transformer and sent it to the lucky guy and he emailed that the
> primary was ok but his digital meter just bounced on the secondary and
> never settled on a reading. I thought maybe I made a bad solder
> connection. The approximate inductance of this secondary is about 360
> henries. He sent it back to me and I measure it as ok, about 1500
> ohms, using my usual methods. I do notice that the reading shows a
> couple of high reading before settling down. This doesn't happen on
> the
> primary which is much lower inductance. My question - are there some
> DMMs, maybe the autoranging types, that have problems with such
> measurements. I could imagine this could happen depending on the rate
> at which it tries to "range" the reading but it would seem that it
> should eventually settle down. He tried it with two different digital
> meters and couldn't get a stable reading. I sometimes use this
> measurement method on 390a parts, hi. Hopefully waiting for
> illumination, Dan.
>
>