In a message dated 8/23/2004 9:09:11 PM Eastern Daylight Time,
DCrespy at aol.com writes:
Dennis,
I assumed you were asking to decide on a substitute? If so: the problem is
not the full scale current, it is the internal resistance of the meter.
Most 1ma movements are 100 ohms. The R-390A needs (if I remember right) 18
ohms. I have successfully used up to 38 ohm movements with completely
comparable results. 100 ohm movements just do not work. If you are
thinking of
shunting a meter, the problem is the same. Most 100 microamp meters are
1000
ohms, so shunted, you still wind up with a 100 ohm / 1 ma meter! The 38 ohm
meters I have used (International brand) had a series internal resistor,
that
when removed dropped the meter resistance.
Good luck and have fun! 73
Harry KG5LO
Saline MI
Very good points Harry! I had investigated using other meters in the past. I
had determined that if you picked up a nice assortment of 100 and 50 microamp
meters (usually at hamfests - 1 3/4" square size) you would hit on one that
had the right amount of internal resistance that could be shunted and end up
with a 0-1ma movement with a resistance in the 18-40 ohm ballpark. From then on
you would know which scale and type of microammeter to be on the lookout
for. Lafayette Radio used to sell a line of Argonne panel meters that would fit
the R-390A front panel. ( 1 3/4" square size ) I have one R-390A with a
Lafayette/Argonne VU meter that fits just right and looks very nice and appears to
have the right sensitivity. That would be one meter to be on the lookout for to
substitute as a line-level meter. 73 Todd Roberts WD4NGG.